… I am encountering issues with data processing and alignment. I have attached screenshots for your reference.
I have converted my raw files to mzML format in centroid mode for upload and have ensured that the I see detectable peaks through MZmine, but I am uncertain about the best peak alignment parameters given that my dataset contains a limited number of detectable peaks. Additionally, I am unsure about the optimal m/z and RT weight settings for peak alignment to improve feature detection.
I would greatly appreciate any guidance you could provide on resolving this issue, particularly regarding preprocessing recommendations, and alignment settings best suited for MetaboAnalyst 6.0.
The HPLC-Orbitrap is a high resolution instrument, and you should NOT use MatchedFiltered which is mainly designed for low-resolution instrument.
For LC-Orbitrap, our tool can perform automated parameter optimization - just leave the “centWave-auto” (default)
If you want to do manual tuning, start from the default parameters for “centWave-manual” assuming you are very familiar with the instument & experiment
You can read these two publications for more detail: